#16693 - RE: HeartBeat Detector
A good idea is to do reverse-reverse-engineering :)) ...
We are so used to do reverse-engg (tailoring the problem to the solution) that we forget its not a bad idea to try proper engg once in a while :D

->Decide the freq range req'd ( 1 - 200 bpm)
->Decide the max accuracy (e.g. 1 bpm or 0.1 bpm) u may need
->Thus for a resolution of 1 bpm and range of 1-200 bpm, you can have 200 distinct values (so result can be an 8-bit value)

->diff between periods of 1 and 2 Hz = 0.5 sec = 500 msec
->diff between periods of 199 and 200 hz = 0.025 msec
This u need a min time base with resolution of 0.025 msec and a max value of 500 msec
that's one huge range!!! In this case I would suggest a timebase of 5-10 microsec

If crystal freq=4 MHz (configure count timebase for 10 microsec pulses)
->for 200 Hz,count=500 =01F4h (16 bits)
->for 199 Hz,count=502 =01F6h (16 bits)
->for 1 Hz,count=10000=186a0h(32 bits)

The difference between counts at higher frequencies is very narrow, so a smaller time base is preferable

You can then either use look-up tables or 32-bit math to derive the result. However if your accuracy goes to 0.1 bpm or lower, you'll need much smaller timebases (order of .2 microsec)

Kundi

List of 13 messages in thread
TopicAuthorDate
HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      
RE: HeartBeat Detector            01/01/70 00:00      

Back to Subject List