??? 08/24/05 14:45 Read: times |
#100003 - Acceptable limits Responding to: ???'s previous message |
+/- 0.5C is acceptable, but due to interference from local 3 phase inverter speed controllers (without any EMC compliance)
the actual noise induced on the probe cables causes a shift of about +/- 2.0C depending upon inverter speed settings, local earthing (seems to depend on weather) etc.etc. I have tried filtering/averaging etc without success. For various reasons, I cant filter the inverters and I cant change the probe system easily, so the only way is to to change the amount of current through the Pt100 whilst the ADC measures. The only good thing is that I dont have to read the probes very often, once a minute would be fine. Chris |